May 9, 2019 01:24 PM
by Dr Ibrahim
Dear all,
Can anybody help me with the problem of SUV calculation using Radiant. We got the imaging on CD file from our PET CT. In about 30% of the cases I can not calculate the SUV. I`m using the CT and AC PET data, always in the same way. The clinic provides needed information in all patients.
Need advise on that.
Thanks
May 9, 2019 07:46 PM
by
Make sure that the Corrected Image (0x0028,0x0051) tag is present in your PET DICOM files.
Also, when the PET series time is later than the acquisition time, it suggests that series could be manipulated (and thus is not original). RadiAnt will not calculate SUV if a series is suspected of manipulation.
May 10, 2019 07:31 AM
by Dr Ibrahim
Thanks for your reply.