Jun 26, 2019 11:35 AM
by DrRomeo79
Hi, I am experiencing issues with the T/I curve since last update. I use it to evaluate breast enhancement kinetics and it has worked flawlessly until now. I put the ellipse roi on the area I mean to study, duplicate the ROI onto other panels, generate curve, confirm and...an error occurs and a message pops up saying that you have to put the ROI onto at least two dynamic sequences. I put it onto 6 or seven dynamic sequences so whats'wrong?
Thanks
Jun 26, 2019 11:59 AM
by DrRomeo79
The MRI machine is a SIEMENS Aera
Jun 29, 2019 05:17 PM
by K.B. Loh
I encountered similar problem with multi-phasic CT liver dataset also since ver 5.0.
Jun 29, 2019 05:37 PM
by
We will release the 5.0.1 version that should fix this issue (most probably today).
Jul 6, 2019 07:46 AM
by K.B. Loh
Thank you for releasing ver 5.0.1.
The bug of Time-intensity Curve solved.
Well done!
Jul 11, 2019 06:45 PM
by
Thank you for confirming this!
Jul 23, 2019 10:58 AM
by DrHoxbroe
Could it be that this problem has not been fixed in the 5.0.1 evaluation version?
Jul 23, 2019 02:27 PM
by
If the TIC tool does not work for you in v. 5.0.1 then it most probably never worked for you... Please contact us at
https://www.radiantviewer.com/contact/ and we will try to find out the reason of your problems.